    liamhuynh
    Participant

    Hi all,

    I run 4 chi.bio reactors with E. coli in 20 mL LB at 25-37 °C, about 24 h per run.

    – Problems

    Stirrers stop during runs (17 stops in 5 runs). Sometimes several reactors stop at the same time, then restart together hours later without anyone touching them. The logs have no gaps, so the computer never lost power.
    One position fails in almost every run, even with a new reactor unit and stirring raised to 1.0.

    – Tried so far

    Stirring 0.6 → 1.0
    New reactor unit in the failing position, and later a second unit
    Shuffling vessels and stir bars between reactors
    Swapping extension cords

    – Questions

    Has anyone seen fans stop and restart by themselves like this? Power supply or connector?
    How can I test one port or cable, e.g. what voltage should the stirrer line have?
    Is one PSU enough for 4 reactors with heaters on and stirring at 1.0?

    Also, my reactor OD usually reads lower than flask OD600. Should I calibrate each reactor against a spectrophotometer?

    Happy to share data. Thanks!

    #2009
    harrison
    Keymaster

    Hey
    THats quite odd, sometimes we had reactors that systematically stop but not multiple in sync. Are there large bits of (magnetic) metal near the reactors that might mess with it?
    Have you tried other stir bars – eg. slightly smaller ones than recommended if it is indeed not working?
    The fact that they behave TOGETHER seems very unexpected and somewhat implies that an external factor is at play – how often have you observed the synchronised aspect of the behaviour?

Viewing 2 posts - 1 through 2 (of 2 total)