Hi all,
I run 4 chi.bio reactors with E. coli in 20 mL LB at 25-37 °C, about 24 h per run.
– Problems
Stirrers stop during runs (17 stops in 5 runs). Sometimes several reactors stop at the same time, then restart together hours later without anyone touching them. The logs have no gaps, so the computer never lost power.
One position fails in almost every run, even with a new reactor unit and stirring raised to 1.0.
– Tried so far
Stirring 0.6 → 1.0
New reactor unit in the failing position, and later a second unit
Shuffling vessels and stir bars between reactors
Swapping extension cords
– Questions
Has anyone seen fans stop and restart by themselves like this? Power supply or connector?
How can I test one port or cable, e.g. what voltage should the stirrer line have?
Is one PSU enough for 4 reactors with heaters on and stirring at 1.0?
Also, my reactor OD usually reads lower than flask OD600. Should I calibrate each reactor against a spectrophotometer?
Happy to share data. Thanks!