    Hi. Sorry if this question has been answered elsewhere but I could not find a clear response.

    The fluorescence that chi.bio measures is somehow normalized by OD? In other words, the output is a fluorescence/cell value or the overall fluorescent signal of the culture? Specifically, we are measuring the fluorescence of an mCherry constitutively expressed in E. coli (starting OD = 0.1) and we are letting the culture grow from OD=0.1 to OD=1. Should we expect the signal to increase together with OD (as in the case of a measured cumulative fluorescence) or should we expect the signal to remain constant (as in the case of measured fluorescence/cell)? Also, the emit FP2 values are in the range of 1.6 while in most of the documentation of chi.bio typical value are in the order of 0.01 or even lower. Is this indicative of something strange?

    Thanks in advance for your assistance.
    best
    m.

    #1763
    harrison
    Keymaster

    Hello Marco,

    Yes, if you see the supplementary of the PLOS Biology paper it details how the “normalisation” works. Basically it is looking at the amount of scattered (excitation) light and dividing the emission light by this to try to mitigate the OD effect. This reduces (by say ~90%) the influence of OD but it is not perfect, so the fluorescence vs cell at OD 0.5 vs 0.1 may be somewhat different.
    In terms of the large FP2/X values it is entirely dependent on what the excitation and emission wavelength bands are. It sounds like your emission wavelength band is not far enough away (e.g. not high enough) compared to the excitation wavelength, hence loads of the excitation light is getting through and this will mean a very poor signal-to-noise ratio in the measurement.

Viewing 2 posts - 1 through 2 (of 2 total)
  • You must be logged in to reply to this topic.
Log in/Register
Scroll to top