  • #1851
    harrison
    Keymaster

    Hello,
    If you are putting in a clean test tube and it is still jumping around a lot, it might be that the reactor is faulty, i.e. the light source isnt on properly.
    WHat is the raw OD0 readings you typically see? If these are <500, likely reactor has a defect. You could ask Labmaker about possibility of fixing this.

    Harrison

    #1859
    Hamed
    Participant

    The raw OD0 is around 13000.

Viewing 2 posts - 1 through 2 (of 2 total)