  • #1547
    uthsc
    Participant

    Do these errors look consistent with liquid-induced failure inside the reactor? This happened for the first time after months without issue. Please let me know if you need additional information. Thank you.

    https://postimg.cc/4K9y4z2d
    https://postimg.cc/k6P42BLy
    https://postimg.cc/R610KvBx

    • This topic was modified 2 years, 4 months ago by uthsc.
    • This topic was modified 2 years, 4 months ago by uthsc.
    #1550
    harrison
    Keymaster

    Hello!
    The errors are consistent with some kind of noise/misadventure on the I2C digital communications. It could be moisture sensor partially triggered, or something else. What I’d recommend is taking some ethanol and wiping down the moisture sensing lines, and then also using either ethanol or an air duster aerosol can to clean the USB connectors on all of the reactor/control computer/pump board (if you put ethanol in them, plug/unplug the USB 10 times, then leave them to dry that would probably do it). The most likely explanation for why the system would (at some random timepoint) STOP working is that there is mess/contamination on those connectors which is leading to those signals getting lost.

    #1551
    uthsc
    Participant

    Thank you very much for the reply and guidance. We will try this over the weekend and report back.